I think this will illustrate what I'm asking:
 

Attachments

  • Test setup.PNG
    Test setup.PNG
    7.2 KB · Views: 536
Disabled Account
Joined 2012
comparison A-P vs QA400

Using the ShibaSoku AG16 source (600 ohms/.7v rms), the QA400 added .0001% THD+N when applied in parallel with Audio-Precision 2722.

That is when the QA400 is in data acquisition mode... when halted and you have only the screen capture display, the THD+N is 2-3%.

The AG16 shows on the A-P as only having 2H and 3H at about -120dB and nothing above the noise floor of -135dB above those two harmonics.

The QA400 however, shows a lot of harmonics and extrainious freqs beyond the first two harmonics and shows the 2H and 3h at a higher level than the A-P.

RNM


DSC01190.jpg

A-P with Shibasoku AG16.jpg

QA400 with ShibaSoku Ag16.jpg
 
Last edited:
Member
Joined 2004
Paid Member
I saw the input induced distortion as well. The interface reduces it but has a similar problem of its own. However the interface does reduce the extraneous spurs.

Adding a buffer/driver to replace the one on the board could work but its still more stuff to deal with. That only makes sense when the interface is inside the box. Connecting flying leads to holes vacated by the BNC connectors is easy. Flying wired to pads on surface mount PCB's demands an additional level of skill. Also, ideal would be an integrated diff in to diff out solution like the one I'm using. Otherwise it will need a number of .1% resistors to have decent common mode rejection.
 
The point being that the QA400 etc. will introduce distortion at its input. You can measure the distortion introduced by the input of the QA400 (or whatever) with this test.

This is interesting. It suggests that one might want to precede the QA400 with a buffer that provides it with a low-Z source impedance. Of course, while doing this one might as well build in an attenuator/protection interface for measuring power amplifiers.

This also makes me wonder what kind of circuitry there is in the QA400 input circuit.

Cheers,
Bob
 
There is a number of interesting op amps designed for low voltage single ended supply with pretty impressive spects. I haven't given them much attention because I like to operate with split supplies at higher voltages. More of these op amps come to market each year with the need for low voltage battery operation. I'll bet there is a much better op amp for the QA400 from this line.
 
Disabled Account
Joined 2012
yes there is.

:):cool:

When you measure a 'system' you see things like this... cant see interfacing/interactions by measuring one component at a time individually apart from one-another.

One reason to get the A-P was to be able to do more complex measurements - like multi-tone tests - thru complete systems. I brought this up last year on another forum (JC's). It is the complete system that we hear but don't measure that way.

Thx-RNMarsh
 
Last edited:
Member
Joined 2004
Paid Member
Good idea but I think the effort would be better spent on a higher performance ADC. I'm meeting with AKM next week and will try to get details on their new ADC. The QA400 is more than adequate for most tasks and overcoming things like the grounding etc. will be the easy improvements.

If you are still set up try changing the source Z (on the AP's generator) driving the analyzer and the QA400 to see if lower source impedance helps. On my setup using the 50 Ohm and 600 Ohm source in the KH4400 I did see a significant difference. I need the effect verified before I get involved with addressing it. Then I need to work over the interface board for the same issues.